Default a 60 Series Indicator

Default a 60 Series Indicator

Postby gseadmin » Wed Apr 08, 2009 7:09 am

How do I default my 60 Series indicator?

Re: Default a 60 Series Indicator

Postby gseadmin » Wed Apr 08, 2009 7:10 am

Default Setup

P65001 : Default All
Defaults all setup parameters to factory default values or to embedded applications. Press [ENTER] to initiate factory default. Press [TARE] to toggle through the choices of embedded applications. Press [ENTER] to load the application.

Original factory parameter configuration can be restored by defaulting the indicator at P65001. ALL SETUP PARAMETERS WILL BE RESET TO THE FACTORY DEFAULT! LOAD CELL CALIBRATION, MULTI-SCALE A/D CALIBRATION, CUSTOM PIN #s AND DATABASE INFORMATION WILL BE LOST!

P65002 : Default - Cal (Save Calibration!)
Defaults all setup parameters to factory default values EXCEPT scale parameters P109 to P136 or to embedded applications. Press [ENTER] to initiate factory default. Press [TARE] to toggle through the choices of embedded applications. Press [ENTER] to load the application.

Other Information
The prompt to "Clr? All DB" will only be displayed if a database option is installed and initialized. A newly installed database option that has not been initialized will be initialized after defaulting.

While defaulting, the display will blank for several seconds, then return to P65001. Saving changes will then overwrite the prvious setup with factory default values. If changes are not saved before exiting the setup, the previous setup is retained.

The Default-CAL at P65002 is similar to P65001, except that all scale parameters (P109 - P136) are left unchanged after defaulting. Thus it is not necessary to recalibrate the load cell or reinstall the multi-scale A/D values.