350/355 Master/Remote to a 50 Series Indicator

350/355 Master/Remote to a 50 Series Indicator

Postby gseadmin » Tue Apr 07, 2009 2:02 pm

1. Can a 350/355 be used as a master to a 50 series indicator? If so, how?

2. Can a 350/355 be used as a remote to a 50 series indicator? If so, how?

Re: 350/355 Master/Remote to a 50 Series Indicator

Postby gseadmin » Tue Apr 07, 2009 2:04 pm

1. Can a 350/355 be used as a master to a 50 series indicator? If so, how?

450/550 Series: The 350/355 can be used as a master to a 450 and 550 series indicator.
* On the 450 and 550, set P160 to Remote Display (9992).
* On the 350/355, set P290 (ECHO) to COMM1.

650 Series: The 650 is not capable of acting as a remote display to any indicator.

2. Can a 350/355 be used as a remote to a 50 series indicator? If so, how?

450 Series: If there is an echo parameter at P290, the 350/355 can be used as a remote. This is in 450 series units dated AFTER July 20, 1995. Any units dated previous CANNOT be used in a master/remote configuration and will NOT have a P290.

550 Series: The 350/355 can be used as a remote display to the 550 indicator.
* On the 550, enable P290 to COMM.
* On the 350/355, set P260 to r-dsp.

650 Series: The 350/355 can be used as a remote display to the 650 series indicator.
* On the 650, set P290 to the COMM Port the 350/355 is connected to.
* On the 350/355, set P260 to r-dsp.